Features
Digital monitor of 4 1/2 digits.
Exclusive use for optical transmission method non-contact flange type torque transducer
Specification
Specification name | Specification contents |
---|---|
Bridge power supply | AC20 Vrms Square wave 11.2 kHz±0.2 kHz |
Applicable transducer | Optical transformer method torque transducer |
Input range | 5 kHz ~ 15 kHz |
Output voltage | DC±10 V, Load resistance 2 k-ohm or more |
Sensitivity adjustable range | ±15 %R.O. (to the rated output of the torque transducers) |
Non-linearity (Display and voltage output) |
0.01 %F.S. |
Temperature efficient | Zero point ±0.025 %R.O./10℃, Sensitivity ±0.025 %R.O./10℃ |
Frequency response range | DC ~ 1 kHz (Filter: W/B) Changeable to 1, 10, 30, 50, 100, 300, 500 or 1kHz) |
Display section | Output voltage display 0 ~ ±10.000 ((Green LED), Sampling Approx. 4 times/s |
Remote function | CHECK |
Alarm function | ALARM is output when the positional gap between rotor and stator is occurred, or when the luminescence part of rotor or the light detecting part of stator becomes dirty. (Photocoupler output) |
Operating temp. / humidity range | 0 ℃ ~ 40 ℃, 85 %RH orless (Non condensing) |
Power supply | 100 VAC±10 V 50/60 Hz, |
Power consumption | Approx.100 VA (at 100 VAC) |
Outline dimensions (WxHxD) | 66 mm x 160 mm x 280 mm (Excludesprotruding parts.) |
Weight | Approx. 2.5 kg |